[sharp-discuss] autosharp not reading column headers in mtz files
Phil Evans
pre at mrc-lmb.cam.ac.uk
Fri Aug 17 12:04:52 CEST 2012
further comments, not immediately relevant
On 17 Aug 2012, at 10:26, Clemens Vonrhein wrote:
> Hi Phil et al,
> ...
> ------------------------------------------------------------------
>
> The easiest solution is to not using MTZ files in autoSHARP but rather
> *.sca files. Those can easily be written out by SCALA or AIMLESS in
> addition to the MTZ files. Just add
>
> OUTPUT POLISH # for SCALA
>
> or
>
> OUTPUT SCALEPACK # for AIMLESS (does POLISH also work?)
>
> The advantage of using *.sca files in autoSHARP:
>
> * SHELXD can read them directly.
>
> So we don't need to convert back from F/SIGF (in MTZ) to I/SIGI (in
> a temporary *.sca file) with some subtle differences: the F's will
> have gone through some truncating procedure and just squaring them
> doesn't give you back the original intensities (which could be
> negative).
>
The MTZ files will normally contain I+ & I- as well as F+ & F-. Do you use Is or Fs (which ShelX needs)? If you read Is from .sca file, how do you convert it to F?
> * you save yourself some typing: no need to give column labels in the
> autoSHARP input form
>
But you do need to select multiple files for MAD. Also autoSharp _could_ make intelligent guesses for the columns :-)
> Some additional tips that could be relevant to this:
>
> * make sure you have the correct scalepack format: the 'merged'
> version is known to work correctly - see
>
> http://www.globalphasing.com/sharp/manual/autoSHARP1.html#datafiles
>
> * make sure you have anomalous data in them (if it is required for
> SIRAS/MIRAS or SAD/MAD)
>
> * it is very helpful to give your MAD wavelengths good identifiers:
> infl, hrem, lrem or peak
>
> The reason: SHELXC doesn't use f'/f" values to distinguish the
> different wavelengths and their corresponding atomic scattering
> properties. It wants names like "HREM", "LREM", "PEAK" or
> "INFL". autoSHARP tries to figure these names out from the f'/f"
> values that you give in the interface - but this can often be
> wrong. So by giving those identifier names, you are helping
> autoSHARP to tell SHELXC the right names. See also
>
> http://www.globalphasing.com/sharp/manual/autoSHARP2.html#iden
>
> Of course, you should still give the correct f'/f" values
> (especially for wavelengths close to the edge) - from a good
> fluorescence scan.
>
The MTZ file has the advantage of including the wavelength, though you still need f'/f" near the edge